:LTESt:MEASure:JITTer:LOCation:UPACtion
Removed command for revision A.03.00 and above. This command is no longer available. Use :LTESt:MEASure:MLIMit:UPACtion
insead.
Command Syntax
:LTESt:MEASure:JITTer:LOCation{1:32}:UPACtion {FAIL | PASS | IGNore}
Query Syntax
:LTESt:MEASure:JITTer:LOCation{1:32}:UPACtion?
Description
Selects an action to be taken if the jitter measurement limit test in Jitter/Noise mode cannot be performed. The jitter measurement is identified by a number, from 1 to 32, that is appended to the :LOCation
portion of the command. Refer to the :MEASure:JITTer:LOCation:STATe command to learn how to identify the measurement and to turn the measurement limit test on or off.
- FAIL
- If the measurement cannot be performed, count it as a failure.
- PASS
- If the measurement cannot be performed, pass the measurement and do not count it as a failure. The total number of waveforms measured is incremented, but the total number of failures is not.
- IGNore
- If the measurement cannot be performed, ignore the measurement and do not count it as a pass or fail.
Use the following sibling commands to configure the measurement:
:FAILures
defines the number of failures for stopping a limit test:FREGion
selects when to fail a measurement:LLIMit
defines the lower limit for failing a measurement:ULIMit
defines the upper limit for failing a measurement
FlexDCA must be in Jitter/Noise mode before you can configure or enable jitter measurement limit tests.