:MEASure:EYE:EECQ:SOURce

Instrument
N1000A
N109x
Flex Apps:
FlexDCA
Meas. mode:
Eye
Waveform type:
PAM
Electrical
Package License:
L-RND

Command Syntax

:MEASure:EYE:EECQ:SOURce {FUNCtion{1:16} | CHANnel{1:8}{A|B|C|D} | DIFF{1:8}{A|C} | CGMemory{1:8} | WMEMory{1:8}}

Query Syntax

:MEASure:EYE:EECQ:SOURce?

Description

Selects the source waveform on which to perform the EECQ measurement. For an example of the command sequence needed to perform the EECQ measurement, refer to the :MEASure:EYE:EECQ command.

The source waveform for the EECQ measurement is meant to be the output of the Reference Receiver waveform signal processing operator (:FUNCtion1:FOPerator RRX). However, you can select any valid, available source waveform.

Requires FlexDCA revision A.07.90 and above.

Example Command

:MEASure:EYE:EECQ:SOURce FUNCtion1