:CALibrate:MODule:PTIMebase:STATus?

Flex Apps:
FlexOTO
FlexOTO Instance:
Hardware Diagram
Instruments:
N1093A/B-PTB

Query Syntax

:CALibrate:MODule:PTIMebase:STATus? "<DCA-M-module-name>"

Where <DCA-M-module-name> is the name that identifies an N1093A/B DCA-M module on the Hardware Diagram. The name is the combination of the <model number>-<serial number>. For example, "N1093B-US12345678".

Description

Returns the status of an N1093A/B DCA-M module's precision timebase calibration. One of the following response strings is returned:

"CALIBRATED"
The calibration is valid.
"UNCALIBRATED"
The calibration is needed.
""
An empty string is returned if the hardware for the indicated slot does not exist. An -241,"Hardware missing" error is returned by the :SYStem:ERRor:Next command.

Use the child queries listed in the following table to return details about the current calibration.

Status String Child Queries
Status String Child Query to Send
Time of last calibration,
Time change since last calibration, and
Temperature change since last calibration
:CALibrate:MODule:PTIMebase:DETails?
Change in temperature :CALibrate:MODule:PTIMebase:DTEMperature?
Change in Time :CALibrate:MODule:PTIMebase:TIME?

Requires FlexOTO revision A.07.90 and above.