:CALibrate:MODule:PTIMebase

Flex Apps:
FlexOTO
FlexOTO Instance:
Hardware Diagram
Instruments:
N1093A/B-PTB

Command Syntax

:CALibrate:MODule:PTIMebase {ENABled | DISabled | PARTial} "<DCA-M-module-name>"

Where <DCA-M-module-name> is the name that identifies an N1093A/B DCA-M module on the Hardware Diagram. For example, "N1093B-US12345678".

Query Syntax

:CALibrate:MODule:PTIMebase? "<DCA-M-module-name>"

Query Response

{ENAB | DIS | PART}

Description

Enables or disables the precision timebase calibration on N1093A/B module.

ENABled
Selects all precision timebase calibrations for the module to be performed.
DISabled
Prevents all precision timebase calibrations for the module from being performed.
PARTial
Indicates that some precision timebase calibrations are enabled and some are disabled.

To start the calibration, use the :CALibrate:MODule:STARt command.

Use the child queries listed in the following table to return details about the current calibration.

Status String Child Queries
Status String Child Query to Send
State of Calibration :STATus?
Time of last calibration,
Time change since last calibration, and
Temperature change since last calibration
:STATus:DETails?
Change in temperature :STATus:DTEMperature?
Change in Time :STATus:TIME?

Requires FlexOTO revision A.07.90 and above.