:CALibrate:MODule:UTIMebase
Command Syntax
:CALibrate:MODule:UTIMebase {ENABled | DISabled | PARTial} "<DCA-M-module-name>"
Where <DCA-M-module-name> is the name that identifies an N1093A/B DCA-M module on the Hardware Diagram. For example, "N1093B-US12345678".
Description
For a indicated N1093A/B module, enables or disables an N1093A/B User Timebase Calibration. The following table lists the input clock signal requirements for the calibration waveform. The source of the calibration waveform must be a calibrated microwave or vector signal generator. Do not use a pattern generator or arbitrary waveform generator. When the calibration starts, the N1093A/B automatically switches to oscilloscope mode.
Connect the signal to an N1093A/B's optical Channel A input and to the Clock In
input.
DCA-M | Frequency | Amplitude |
---|---|---|
N109x-series | 26.000000 GHz | 100 mVpp to 700 mVpp |
Requires FlexOTO revision A.07.90 and above.
Example Command Sequence
These steps are used for an N1093A/B timebase calibration in slot 5.
:CALibrate:MODule:UTIMebase "N1093B-US12345678" :CALibrate:SDONe? … // Prompt user to disconnect all cables to N1090A. :CALibrate:CONTinue :CALibrate:SDONe? … // Prompt user to connect calibration sinewave. :CALibrate:CONTinue …