:CALibrate:SDONe?

Flex Apps:
FlexDCA
FlexRT
FlexPLL
FlexOTO

Query Syntax

:CALibrate:SDONe?

Query Response

{1 | 0}

Description

Queries the completion of a calibration step, which results in a dialog being displayed where the user must choose either to continue or cancel the calibration. The CAL:SDONe? query response string is the message shown in the dialog enclosed in double quote characters and followed by a newline character.

For example, following string is returned during a mainframe user timebase calibration when the introductory dialog is displayed:

"Please connect a 10.000000 GHz clock signal to the Calibration Channel and to the Front Panel Trigger. The amplitude must be between 100 mV p-p and 700 mV p-p. Important: You must use a calibrated microwave or vector signal generator. Do not use a pattern generator or arbitrary waveform generator."\n.

At this point, the user clicks a NEXT or CANCEL button, which is accomplished remotely using the :CALibrate:CONTinue or :CALibrate:CANCel commands.

If a calibration has not been initiated, the following string is returned.

"No Calibration in progress."

Example Command Sequence

The following commands are used for a typical vertical calibration.

:CALibrate:SLOT3:VERTical:STARt
:CALibrate:SDONe?
:CALibrate:CONTinue    //  Dismiss the begin notice calibration dialog box
:CALibrate:SDONe?
:CALibrate:CONTinue    //  Dismiss the disconnect notice dialog box
:CALibrate:SDONe?
:CALibrate:CONTinue    //  Dismiss calibration complete dialog box
*OPC?