:CALibrate:SLOT:STATus:DETails?

Flex Apps:
FlexDCA
FlexRT

Query Syntax

:CALibrate:SLOTN:STATus:DETails?

Where N identifies a specific module by slot and must be selected from {1:8}. For a module that occupies multiple slots, use the module's left-most slot number to identify the module. For two-slot modules, the slot number would be 1 or 3. For four-slot modules like the 86108B, the slot number would be 1.

Description

Queries a string of status details for all calibrations for the indicated slot (module). The following is an example of a typical returned string:

Time of last Calibration: 10/23/2012 11:52:00,Elapsed Calibration Time: 3 hours 2 minutes,Temperature Change: 1.24 C"

If a calibration is recommended, the following text is placed at the beginning of the above string:

"Requires Clock Recovery Calibration, …"

If a calibration has not yet been performed, the following string is returned:

"This calibration has never been performed in this instrument."

Example Returned Status String

"Requires Clock Recovery Calibration,Time of last Calibration: 03/14/2012 11:52:00,Elapsed Calibration Time: 75 hours 2 minutes,Temperature Change: 3.40 C"

Example Returned Status String

"Requires Vertical Calibration.,Requires Precision Timebase and Clock Recovery Calibration.,Time of last Calibration: 08/03/2011 13:19:02,Elapsed Calibration Time: 355 hours 15 minutes,Temperature Change: 11.4°C"

Parent Command