:MEASure:JITTer:OJITter:JRMS
Command Syntax
:MEASure:JITTer:OJITter:JRMS
Query Syntax
:MEASure:JITTer:OJITter:JRMS?
Description
Send this command to move the Jrms IEEE output jitter measurement results to the top of Jitter Mode's Output Jitter panel. Send the query form of this command to return the IEEE JRMS measurement result. The default returned measurement is based on all waveform edges. To return a measurement result that is based on one of twelve specific waveform edges, use the :MEASure:JITTer:OJITter:JRMS:ECATegory
command.
Before the :MEASure:JITTer:OJITter:JRMS
command can be used, you must turn on all IEEE measurements by sending the :MEASure:JITTer:OJITter:STATe
command.
Measurement Identification
Avoid subtle programming errors! To ensure that the correct measurement is installed or queried, always explicitly identify a measurement when installing a measurement or querying a measured value, status, or detail. To identify a measurement, specify the measurement's source waveform (:SOURce
child command). With some measurements, you may also need to specify other identifying values. Generally, when selecting a measurement using FlexDCA's GUI, if a dialog appears prompting you to select values, you should explicitly specify these values when remotely identifying the measurement. More information.
For example, to return the status of the measurement, :MEASure:JITTer:OJITter:JRMS
:
flex.write(':MEASure:JITTer:OJITter:JRMS:SOURce CHAN1A') if flex.query(':MEASure:JITTer:OJITter:JRMS:STATus?') == 'CORR'; measurement = flex.query(':MEASure:JITTer:OJITter:JRMS?') else: details = flex.query(':MEASure:JITTer:OJITter:JRMS:STATus:DETails?') reason = flex.query(':MEASure:JITTer:OJITter:JRMS:STATus:REASon?')
Child commands for measurement identification:
:MEASure:JITTer:OJITter:JRMS:SOURce
Requires FlexDCA revision A.05.70 and above.
Measurement Ready?
To confirm that the measurement is ready to read, you can query the measurement's status:
if ('CORR' in Flex.query(':MEASure:JITTer:OJITter:JRMS:STATus?')): measurement = Flex.query(':MEASure:JITTer:OJITter:JRMS?')
If averaging is turned on (:ACQuire:AVERaging
), you can also confirm if the measurement result is ready by comparing the number of specified averages (:ACQuire:ECOunt?
) versus the number of measurement sweeps that have occurred (:MEASure:JITTer:OJITter:JRMS:COUNt?
). The technique uses the :COUNt?
common measurement query. For example,
if (Flex.query(':MEASure:JITTer:OJITter:JRMS:COUNt?') >= Flex.query(':ACQuire:ECOUNt?')): measurement = Flex.query(':MEASure:JITTer:OJITter:JRMS?')
You can also use an acquisition limit lest to test that a number of waveform samples, or pattern acquisitions have completed before returning a measurement. Refer to the :LTESt:ACQuire:CTYPe
command.
Common Measurement Child Queries
This command supports the use of the following common measurement queries: :LOCation?
, :COUNt?
, :MAXimum?
, :MINimum?
, :MEAN?
, and :SDEViation?
.