N1000A DC Cal Output
Instrument
N1000A
Flex Apps:
FlexDCA
The N1000A's front-panel calibration output (DC Cal) is used to manually verify the DC accuracy of a module's electrical channel calibration.
In the Front Panel Cal Output dialog, you can adjust the DC Cal output voltage to 0V ±2V. For best calibration accuracy, set the calibration output level to a voltage other than 0.0V. The default level is 5.0 mV. Setting the level above approximately ±800 mV will require that you offset the waveform display, add an external attenuator, or both. The equivalent SCPI command is :CALibrate:OUTPut.
To Verify Electrical-Channel Calibration DC Accuracy
- Connect a cable between the DC Cal output and the module's channel input. Or, use the DC Cal output probe hook.
- Set the calibration output level using Front Panel Cal Output dialog.
- Confirm that the measured calibration output level is the level entered in the dialog.