Skew Calibration
The skew calibration removes any difference in the electrical lengths between a module's two remote sampling head cables.
Calibration commands correspond to the features in the Calibration dialog. You can click on the tabs in this picture to learn about the different calibrations.
![](../../../Resources/Images/FlexDCA/calibration_skew_tab.png)
Use the :CALibrate:SKEW:SLOT:STARt
command to start a skew calibration for a specific slot. You do not select or enable a skew calibration as you do with other calibrations. To query the status of a skew calibration, use the :SKEW:SLOT:STATus?
command.
The skew calibration is can be performed on N1000-series plug-in modules (for example, N1045A), a DCA-M module (for example, N1094A/B), or an 86118A-H01 module.
To run the calibration
- Connect a 1.0 GHz sine wave to one of the input channels.
- Connect a synchronous trigger to the DCA-X's front-panel Trigger input. The amplitude of the signal must be between 200 mV (-10.0 dBm) and 710 mV (1.0 dBm).
- Send the
:CALibrate:SKEW:SLOT:START
command to start the calibration.
During the calibration, always use same cables and connectors to connect the source to each of the module's channels.
Do not remove a module from the DCA-X during a calibration.
The input circuits can be damaged by electrostatic discharge (ESD). Therefore, avoid applying static discharges to the front-panel input connectors. Prior to connecting any coaxial cable to the connectors, momentarily short the center and outer conductors of the cable together. Avoid touching the front-panel input connectors without first touching the frame of the instrument. Be sure that the instrument is properly earth-grounded to prevent buildup of static charge. An anti-static mat and wristband are strongly recommended.