:CHANnel:RTSCope:INTerpolate:FACTor

Instrument:
UXR Scope
Meas. mode:
Scope
Eye
Flex Apps:
FlexRT

Command Syntax

:CHANnel{1:8}{A|B|C|D}:RTSCope:INTerpolate:FACTor <factor>

Where <factor> is an integer.

Where the required specifier {1:8}{A|B|C|D} identifies a channel by its module slot number and channel letter. For example, channel 3C (CHANnel3C).

The channel specifier will fail if it does not include a slot number. If instead the suffix's channel letter is missing, the command defaults to letter A. As a result, the best practice is to always explicitly specify the channel by including both the module slot number and channel letter.

Query Syntax

:CHANnel{1:8}{A|B|C|D}:RTSCope:INTerpolate:FACTor?

Description

For the selected channel, specifies a factor (integer value) that determines the number of points FlexRT uses to interpolate between points on the real-time oscilloscope's waveform. To select the interpolation method, use the :CHANnel:RTScope:INTerpolate:METHod command. The number of sample-per-UI in the interpolated waveform is:

(acquired samples-per-UI) (factor) = samples-per-UI

You can also directly set the sample-per-UI using the :CHANnel:RTScope:INTerpolate:SPUI command.

Before this command can take effect, you must disable shared acquisition parameters with the :CHANnel:RTSCope:TRACkcommand.

Requires FlexDCA revision A.06.70 and above.

Example Command

:CHANnel2A:RTSCope:INTerpolate:FACTor 2