86100D Horizontal (Timebase) Specifications
Instrument:
86100D
Specifications describe warranted performance. Characteristics provide useful, nonwarranted information about the functions and performance of the instrument. Characteristics are printed in green italics.
Item | Description |
---|---|
Scale Factor | Full scale is ten divisions. |
Minimum | 100 fs/div |
Maximum | 1 s/div (250 ns/div with pattern lock) |
Delay | Time offset relative to the front panel trigger input on the instrument mainframe. |
Minimum | 24 ns 40.1 ns default, 24 ns minimum (with Pattern Lock) |
Maximum | 1000 screen diameters or 10s, whichever is smaller 1000 screen diameters or 2.5401 μs, whichever is smaller (with Pattern Lock) |
Time Interval Accuracy Dual marker measurement performed at a temperature within ±5° C of horizontal calibration temperature. |
1 ps + 1.0% of Δ time interval or 8 ps, whichever is smaller. The maximum delay settings is = 100 ns and delta time does not span across (28 + N x 4) ns ±100 ps delay setting, where N = 0, 1, 2, 18. If delta time measurement span exceeds this criteria, time interval accuracy is 8 ps + 0.1% of Δ time reading. |
Jitter Mode Operation | Time interval accuracy – jitter mode operation 1 ps. Characteristic. Test configuration: PRBS of length 27–1 bits, Data and Clock 10 Gb/s. |
Time Interval Resolution. The time interval resolution is the smallest time you can characterize between two points. | ≤ (screen diameter)/(record length) or 62.5 fs, whichever is larger |
Display Units | Bits or Time |
Record length | 16 to 65,536 samples. Increments of 1 |