Probes

Introduction

On an N1000A or 86100D, you can connect probes to a module's channel inputs. On 54752A, 54754A, 83484A, 83486A, 86108A, 86108B, and 86112A modules, a probe power connector, shown in the following figure, allows an N1000A or 86100D to automatically detect when a probe-power connection to a probe is made. Because information about the attached probe cannot be detected, you must use the External Hardware Setup dialog to enter the probes parameters.

When you connect the adapter's power cable to the module's front-panel probe power connector, the Probe Setup dialog is displayed. In this dialog, select the probe that most closely matches the probe you are using.

Keysight Probes that are directly supported are listed in the following tables. If your probe is:

  • Not a directly supported model but still requires an N1022A/B or N5477A adapter, connect the probe and specify a model number of a similar probe that has the same attenuation factor.
  • Not similar to a supported probe or does not use an adapter, define the probe as an external attenuator in the External Hardware Setup dialog.

The following procedures are provided for connecting probes:

After a probe is connected, perform a Probe calibration.

Due to a wide variety of signals that may be encountered, which range from high bandwidth (fast rise times) to high impedance, one type of probe may be more effective than another type. Therefore, it is helpful to understand the different effects caused by the interaction between the probed circuit and the probe. The response of the probe is part of the overall measurement system response. When a probe is connected to the instrument to measure a signal it becomes part of the circuit under test. The probe can then degrade the circuit under test. The new circuit will behave differently than the circuit without the probe. The following picture displays the probe as part of the circuit under test.

Directly Supported Probes

Refer to the individual probe documentation for the latest information on product specifications.

Always use the proper connection technique when connecting N2800A-series probes to the oscilloscope.

Directly Supported InfiniiMax III Probes
Probe Attenuation
Factor
Bandwidth ZIn Dynamic Range Required
Adapter
N2803A 6:1 30 GHz 100 KΩ 2.5 Vp-p N5477A
N2802A 6:1 25 GHz 100 KΩ 2.5 Vp-p
N2801A 6:1 20 GHz 100 KΩ 2.5 Vp-p
N2800A 6:1 16 GHz 100 KΩ 2.5 Vp-p
Directly Supported InfiniiMax I & II Probes
Probe Attenuation
Factor
Bandwidth Dynamic Range DC Offset Range Input Required
Adapter
1169A 3.45:1 13 GHz (typical) ± 3.3 V ± 16 V Differential Input R: 50K ohm
Differential Input C: 0.21 pF.
Single Ended Input R: 25K ohm.
Single Ended Input C: 0.35 pF.
N1022A/B
1168A 3.45:1 10 GHz 3.3 V pk-pk ± 16 V Differential Input R: 50K ohm. Differential Input C: 0.21 pF.
Single Ended Input R: 25K ohm. Single Ended Input C: 0.35 pF.
1158A 10:1 4 GHz ± 5V DC ± 15V DC
1157A 10:1 2.5 GHz ± 5V DC ± 15V DC
1156A 10:1 1.5 GHz ± 5V DC ± 15V DC
1152A 10:1 2.5 GHz
1134A 10:1 7 GHz ± 2.5 Vdc ± 12V 25 KΩ single mode
50 KΩ differential mode input resistance
1132A 10:1 5 GHz ± 2.5 Vdc ± 12V 25 KΩ single mode
50 KΩ differential mode input resistance
1131A 10:1 3.5 GHz ± 2.5 25 KΩ single mode
50 KΩ differential mode input resistance