:CALibrate:SLOT:CRECovery
Command Syntax
:CALibrate:SLOTN:CRECovery {ENABled | DISabled}
{N} identifies a specific module by slot and must be selected from {1:8}. For a module that occupies multiple slots, use the module's left-most slot number to identify the module. For two-slot modules, the slot number would be 1 or 3. For four-slot modules, like the 86108B, the slot number would be 1.
Query Syntax
:CALibrate:SLOTN:CRECovery?
Query Response
{ENAB | DIS}
Description
For the indicated slot (slot), enables or disables the clock recovery calibration. To perform this calibration along with all other selected calibrations for the indicated slot, use the :CALibrate:SLOT:STARt
command. To select a vertical or precision timebase calibration, use the CALibrate:SLOTN:VERTical
and CALibrate:SLOTN:PTIMebase
commands. To select all calibrations for the indicated slot, use the CALibrate:SLOTN:ENABled
command. These commands affect the Calibration dialog indicators, as shown in the following picture.
- ENABled
- Selects the calibration.
- DISabled
- Prevents the calibration.
Requires FlexDCA revision A.01.70 and above.
Use the child queries listed in the following table to return details about the current calibration.
Status Sring | Child Query to Send |
---|---|
State of Calibration | :STATus?
|
Time of last calibration, Time change since last calibration, and Temperature change since last calibration |
:DETails?
|
Change in temperature | :DTEMperature?
|
Change in Time | :TIME?
|
Child Commands
:STATus?
queries the measurement status.:STATus:DETails
:STATus:DTEMperature
:STATus:TIME