:CALibrate:TIMebase:SLOT:STARt
Command Syntax
:CALibrate:TIMebase:SLOTN:STARt
Where N identifies a specific DCA-M N109x-series module by slot and must be selected from {1:8}.
Description
For a DCA-M module in the indicated slot, starts a N109x-series Timebase Calibration. The following table lists the input clock signal requirements for the calibration waveform. The source of the calibration waveform must be a calibrated microwave or vector signal generator. Do not use a pattern generator or arbitrary waveform generator. When the calibration starts, the DCA-M automatically switches to oscilloscope mode.
Connect the signal to an DCA-M's electrical channel input and to the front-panel Clock In
input. Use the :CALibrate:TIMebase:SLOT:CHANnel
command to specify the input channel. After performing the calibration.
DCA-M | Frequency | Amplitude |
---|---|---|
N109x-series | 26.000000 GHz | 100 mVpp to 700 mVpp |
Requires FlexDCA revision A.03.00 and above.
Example Command Sequence
These steps are used for an N1090A timebase calibration in slot 5.
:CALibrate:TIMebase:SLOT5:STARt :CALibrate:SDONe? … // Prompt user to disconnect all cables to N1090A. :CALibrate:CONTinue :CALibrate:SDONe? … // Propt user to connect calibration sinewave. :CALibrate:CONTinue …