TDECQ Operator (Measurement for NRZ)
This topics describes this dialog tab for NRZ waveforms. Click here to learn about this tab for PAM waveforms.
Use the TDECQ Measurement tab of the TDECQ Reference Equalizer Setup dialog to globally configure all TDECQ measurements. The TDECQ measurement settings in this dialog tab are linked to and also set identical settings located in the dialog tabs listed in the following table.
| Button | Feature | Linked Dialog Settings |
|---|---|---|
|
Eye Mode PAM Measurement Configuration |
PAM-N Analysis Setup dialog's TDECQ Configuration tab. |
|
Signal Processing Operators |
TDECQ Reference Equalizer Setup dialog's TDECQ Operator (Measurement for PAM). |
|
Reference Receiver Setup dialog's TDECQ Secondary Operator (Measurement for PAM). |
The NRZ TDEC measurement settings in this dialog tab are the same settings found in the TDECQ Reference Equalizer Setup dialog's TDEC Measurement tab. Change a setting in one of these dialogs and the setting is changed in both places.
Histogram Properties
The histogram windows used in the TDEC and VECP measurements are set to a default width of 0.04 UI. This setting is applied when FlexDCA is not in pattern lock. It adjusts the histogram window so that additional data can be acquired to decrease the measurement time. The window's width can be set from 0.01 UI to 0.20 UI. The default setting is 0.04 UI. The default value is the same value used in 100GBASE-SR4.
| Minimum Width (0.01 UI) | Maximum Width (0.20 UI) |
|---|---|
| TDEC Measurement (four histogram windows) | |
|
|
| VECP Measurement (two histogram windows) | |
|
|
Histogram Position
The position of the left and right histograms can be positioned on the waveform. The position of the left histogram can be set from 0.25 UI to 0.50 UI. The position of the right histogram can be set from 0.50 UI to 0.75 UI.
Levels Definition
The following paragraphs describe how these settings work together. Both the settings shown in the TDECQ operator dialog's TDEC Measurement tab and the associated remote SCPI commands are noted.
Minimum Zero Level and One Level Run Settings
The Minimum Zero Level Run and the Minimum One Level Run settings enter the minimum number of consecutive UIs that must be present in the waveform pattern in order to measure the amplitude of the level. Enter the minimum number of consecutive zero or one UIs that must be present in the waveform. If the runs are shorter, the measurement is marked a questionable (a "?" character annotates the measurement's results). This number includes the actual 2 UIs where the measurement is made plus margin UIs on either side.
| Settings | SCPI Command | Default Values |
|---|---|---|
| Minimum Zero Level Run1 | :MEAS:TDEC:MRUN:ZERO
|
72 |
| Minimum One Level Run1 | :MEAS:TDEC:MRUN:ONE
|
72 |
| Run Measure Length | :MEAS:TDEC:LMLength
|
50 UI |
Setting includes 2 measured UIs plus margin UIs.
For default minimum setting of zero level runs:
For default minimum setting of level three runs:
If the runs present in the pattern are shorter than this setting, the measurement is marked a questionable (a "?" character annotates the measurement's results). The default values are 6 runs of zero and 7 runs of level 3. The values can range from 5 to 1024 UIs.
Longer Runs Use and Location Method Settings
The IEEE 802.3dj Draft 3.2 standard introduced changes to the TDEC/VECP measurement algorithm when performing TDEC measurements. To support this change, the new Longer Runs Use and Location Method settings were added to FlexDCA A.08.40.
When performing measurements, you can select to use either legacy or new IEEE 802.3dj Draft 3.2 settings as shown in the following table. The legacy settings are the default selections. In the dialog's Preset drop down, if you select an IEEE 802.3dj Draft 3.2 preset, the Longer Runs Use and Location Method settings are set to the correct algorithm setting. Selecting any other preset will set these two settings to their default legacy values. Figure 3 illustrates the differences between the Maximum Run and Min Run at Start settings.
The Longer Runs Use setting specifies how to treat runs that are longer than the minimum. The legacy setting Maximum Run uses all data from the runs excluding the margin at the start and end of the run. The new Min Run at Start setting uses only the 2UI immediately after the margin for the measurement. See Figure 3 below.
The Location Method setting determines how level three runs and level zeros runs are located within the waveform. The legacy algorithm was not pattern aware so it estimated the locations of level three and level zero runs by measuring amplitudes. The legacy algorithm can be selected as Estimate From Amplitude. See Figure 1 below. The new Use Pattern setting causes FlexDCA to first detect the symbol pattern based on the input pattern settings. The detected pattern is used to determine the measurement location. See Figure 2 below.
| Settings | SCPI Command | Legacy Default Values (SCPI Argument) |
IEEE 802.3dj Draft 3.2 Default Values (SCPI Argument) |
|---|---|---|---|
| Longer Runs Use | :MEAS:TDEC:LRUSe
|
Maximum Run ( MWIDth) |
Min Run at Start ( MRAStart) |
| Location Method | :MEAS:TDEC:LMEThod
|
Estimate From Amplitude ( EFAMplitude) |
Use Pattern ( UPATtern) |
FlexDCA's legacy measurement algorithm for 0 level and 3 levels, could locate level 3/0 pattern runs based on the waveform value exceeding a lower limit for 3 levels or dropping below an upper limit for 0 levels. Figure 1 and Figure 2 illustrate the two methods.
Presets
Presets allow you to save your dialog settings to a setup file. To save your settings, click the
button. Select a preset in the drop-down list to instantly configure your TDEC setting to your specification. You can save as many presets as you need. The Preset list shows all of the factory provided presets as well as any that you have created.
TDEC Measurement Presets
- 200G OCI MSA
- 400G CWDM8 MSA 10km Optical Interface
- IEEE 802.3bm (100 GbEn)
- ITU 50G PON
- Stressed Eye Closure (SEC)
- T11 FC-PI-6P (32xFibre Channel)
If you scroll to the end of the list and click the <Edit List> entry, the Edit TDEC Presets List dialog opens which allows you to reorder, delete, or rename items in the list. In the list, the names of factory provided presets includes the text (factory) which indicates that the entry cannot be edited.
TDEC
Target BER
The optical power penalty of the measured optical transmitter is based on the amount of noise that would need to be added to obtain a target BER. The default setting is 5 x 10−5 BER, which is called out for 100GBASE-SR4 which uses FEC. The range of values can be from 1 x 10−12 to 1 x 10−1. Target BER is applied to TDEC measurements only and not to VECP measurements.
Modal Noise Factor
Modal Noise Factor is a property of the fiber-optic link that is connected to the output of the optical transmitter DUT. The default setting is 1.75 x 10−2. The range of values can be from 0 to 5 x 10−1. Mode Noise Factor is applied to TDEC measurements only and not to VECP measurements.
Mode Partition Noise Factor
Mode Partition Noise Factor is a property of the fiber-optic link that is connected to the output of the optical transmitter DUT. The default setting is 4.0 x 10−2. The range of values can be from 0 to 5 x 10−1. Mode Partition Noise Factor is applied to TDEC measurements only and not to VECP measurements.
TDEC One Level Noise Gain
Sets the TDEC measurement's One Level Noise Gain which affects TDEC and VECP measurements. The default value is 1.000 and the range is 1.000 to 12.000.