:CALibrate:OPTical:USER:CHANnel:STARt

Flex Apps:
FlexDCA
FlexRT

Command Syntax

:CALibrate:OPTical:USER:CHANnel{1:8}{A|B|C|D}:STARt

Where the required specifier {1:8}{A|B|C|D} identifies a channel by its module slot number and channel letter. For example, channel 3C (CHANnel3C).

The channel specifier will fail if it does not include a slot number. If instead the suffix's channel letter is missing, the command defaults to letter A. As a result, the best practice is to always explicitly specify the channel by including both the module slot number and channel letter.

Description

On N1090A DCA-M extended hardware modules, starts an optical calibration at a user wavelength on the indicated channel.

To specify the wavelength for the calibration, use the :WAVelength sibling command. To specify the optical power used in the calibration, use the :OPOWer sibling command. The wavelength and source power used must be within the ranges listed in the following table.

Optical Calibration Requirements
Module Factory Calibrated Wavelength(s) Allowable Calibration Optical Source Power Level Source Power for Calibration to Meet Specifications
N1090A 850 nm
1310 nm
1550 nm
375 μW to 425 μW
175 μW to 225 μW
175 μW to 225 μW
375 μW to 425 μW
175 μW to 225 μW
175 μW to 225 μW

The minimum power level is recommended to assure specified performance over the instrument's full dynamic range. The calibration accuracy is heavily dependent on the accuracy to which you know the optical source power. For best results measure the optical source power with a high precision optical power meter.

Use the :STATus? sibling query to return details about the current calibration.

Requires FlexDCA revision A.05.00 and above.

Example Command

:CALibrate:OPTical:USER:CHANnel2A:STARt