:SPRocess:RRX:RPOint:IFOutput

Instrument:
N1000A-PLK
N109X-PLK
Flex Apps:
FlexDCA
FlexRT
Meas. mode:
Scope
Eye
Jitter
TDR
Package License:
L-RND

Command Syntax

:SPRocessN:RRX:RPOint:IFOutput {ON | OFF}

Where N identifies one of up to 64 possible operators {1:64} that can be defined as part of a math function. The operators are numbered in the order that they were originally added to the function.

Query Syntax

:SPRocessN:RRX:RPOint:IFOutput?

Description

For the Reference Receiver operator, selects the Infer Reference OOMA from Output setting. This setting causes an Outer Optical Modulation Amplitude (OOMA) measurement to be made on the waveform at the output of the TDECQ operator. When unselected, the measurement is made on the waveform at the input to the TDECQ operator. By default, this setting is ON.

Infers that the reference OOMA from the output of the Reference Receiver operator. This is the default setting (ON). To disable this setting, select OFF.

Requires FlexDCA revision A.08.40 and above.