:MEASure:JITTer:DEFine:LEVel:CUSTom:EYE

Meas. mode:
Jitter
Waveform type:
NRZ
PAM4
Flex Apps:
FlexDCA

Command Syntax

:MEASure:JITTer:DEFine:LEVel:CUSTom:EYE{N} <custom>

{N} is 0 for an NRZ signal and 0, 1, or 2 for a PAM4 signal.

<custom> is a double.

Query Syntax

:MEASure:JITTer:DEFine:LEVel:CUSTom:EYE{N}?

Description

For the specified eye, sets the jitter sampling level at a custom amplitude value. For NRZ signals, this is the value for the single eye. For PAM4 signals, you can individually specify the percentage that is used for each of the three eyes. If you enter a custom level that is outside of the equivalent 30% to 70% range for the signal being measured, the measurement graphs are not displayed.

To enter a custom amplitude, the :MEASure:JITTer:DEFine:LEVel:TYPe command must be set to UNITs.

Requires FlexDCA revision A.05.60 and above. This command replaces deprecated command :MEASure:JITTer:DEFine:LEVel:CUSTom.